07-01-2008, 04:00 PM
:confused:
The left Vol Increments without control and doesnt stop when it is connected to the simulator, it remains ok when its not connected. Also the ABP increments. (Left Vol 425000, ABP 2296/200)

Incrementing 1000 ml/sec(aprox).

thanks

nbrent
07-06-2008, 01:36 AM
I recently had the same problem with mine. I had to send it back to METI, but it works greeat and their cust support is great. Good luck:)

jjacoby
07-07-2008, 05:48 PM
Sorry to hear about the lung problem. On an ECS simulator this can be caused by a disconnected or malfunctioning flow sensor (the ones used for measuring external ventilations). Like Nathan said, if you contact customer support they should be able to diagnose and remedy the situation.